Kepco SNR 488-8 Too Many listners on GPIB

Hello,

 

I am having problems connecting to a Kepco SNR 488-8 Power Controller via the GPIB bus. The odd thing is that I was able to connect to it before...

 

Here are the steps to reproduce the error:

  • Start NI MAX
  • Click on My System -> Device and Interfaces -> PCI-GPIB "GPIB0"
  • On the "Properties" tab, click on the button "Scan for Instruments"
  • Get an error that says "Too many listeners on GPIB"

This is what I've tried to fix the error:

  • I verified that the PCI-GPIB is set to address 0.
  • I also verified that the address for the Kepco SNR 488-8 is address 4.
  • I can't click on "Restore Defaults" button for the GPIB card as it's grayed out.
  • I tried two different GPIB cables connecting directly from the PC to the test equipment, so it's unlikely that it's a bad cable.
  • I updated to the latest version of NI MAX v5.5.0f0.
  • I tried the NI-488.2 troubleshooting utility. Both of the options passed.
  • I also collected a trace if that helps.

I had the Agilent VISA installed at one point, but I uninstalled it so the NI VISA should be the only one left.

 

Is there anything else I can try?

 

Thank you,

Graham Best

 

0 Kudos
Message 1 of 5
(4,294 Views)
It sounds to me that you have a hardware fault with the Instrument or the controller. Is there another instrument that you can connect?

p.s. The version of MAX is not that relevant. What's your version of NI-488.2?
0 Kudos
Message 2 of 5
(4,288 Views)

The version of NI-488.2 is 3.1.1.

 

I was able to connect to another test equipment device. I also tried to connect to another Kepco SNR 488-8 power controller and had the same issue. Both of these units are functional when they are connected to a much older computer.

0 Kudos
Message 3 of 5
(4,238 Views)
You could try slowing down the bus timing. This is a setting of the controller's properties in MAX.
0 Kudos
Message 4 of 5
(4,233 Views)

I tried changing the bus timing from 500 nanoseconds to 2 microseconds and the instrument was still not recognized.

0 Kudos
Message 5 of 5
(4,228 Views)