06-08-2017 09:13 AM
I have Keithley 2601 and I am using attached example vi from NI. But I am getting error 1073807339, and instrument doesn't go into remote state.
06-16-2017 09:02 AM
Hi Amjad_F,
It looks to be a timeout error. Did you already check these resolved discussions?
http://forums.ni.com/ni/board/crawl_message?board.id=140&message.id=40032
Additional the timing troubleshoot could help:
http://digital.ni.com/public.nsf/allkb/874B379E24C0A0D686256FCF007A6EA0
Best regards
Best regards
06-19-2017 07:28 AM
Are you certain that your hardware connection is correct? Can you communicate with the instrument with NI-Max?
Try running this simple vi to communicate with the instrument.The other case '-1' passes the inputs right on through to the outputs. Enter into the write buffer:
*IDN?\n
See what response the instrument gives. The response should begin with 'Keithley..."