01-28-2015 07:59 AM - edited 01-28-2015 08:01 AM
I'm using 2 Keithley 2400 in a measurement setup. The outputs of the devices are being switched on and off a lot of times. Therefore, they are always 'beeping".
How do I programmatically turn of those beeps?
01-28-2015 08:07 AM
What does the programmer's manual say?
01-28-2015 08:32 AM
STATe <b>
:SYSTem:BEEPer:STATe <b> Enable or disable beeper
Parameters <b> = 1 or ON Enable beeper
0 or OFF Disable beeper
Query :STATe? Query state of beeper
Description This command is used to enable or disable the beeper. When enabled, a
short beep is provided to signal that a front panel key has been pressed.
This only works for the front panel keys, but not for the output enabled beep?
01-28-2015 08:48 AM
01-28-2015 02:27 PM
01-28-2015 02:31 PM
Here is the exact command I used to turn off that very annoying beep for the Keithley 2425. You will need to send this command after every reset if you want it to stay off.
:SYST:BEEP:STAT OFF
01-29-2015 03:37 AM
According to the User Manual of the 2400, the command which crossrulz recommends should also work with the 2400
Model 2400 Series SourceMeter User's Manual - Keithley2400Manual.pdf
http://research.physics.illinois.edu/bezryadin/labprotocol/Keithley2400Manual.pdf
Page 399, at the bottom of the page.
01-29-2015 06:12 AM
@Stephan_D wrote:
According to the User Manual of the 2400, the command which crossrulz recommends should also work with the 2400.
It had better. They are in the same series after all.
02-07-2017 08:39 AM - edited 02-07-2017 08:43 AM
:SYST:BEEP:STAT OFF
Thanks mate worked like a charm 🙂
02-17-2017 08:17 AM
Look into the SYSTem subsystem section for the 2400 manual (Table 18-9). The default setting for the SMU is to power-up in RESET mode which turns the Beeper back (ON). This can be changed to any of the other
:SYST:POS <name>
options as well that will allow you to set the Beep (OFF) and keep it off through power cycles.