Getting Yokogawa WT210 to Interact with LabVIEW (or anything for that matter)

Solved!
Go to solution
Solution
Accepted by topic author tomg09

TG,

 

Make sure you have 'Mode' configured for 488.2 on the WT210.  Make sure you have a Null Modem cable.

Message 11 of 16
(3,465 Views)

Thanks Wayne, let me give that a shot.

0 Kudos
Message 12 of 16
(3,464 Views)
Why should mode be 488.2 when a serial connection is desired?
0 Kudos
Message 13 of 16
(3,460 Views)

It's in the manual.  Have a look at Chapter 11.

0 Kudos
Message 14 of 16
(3,457 Views)

It works!  Thanks a lot Wayne!

 

EDIT: If anyone wants to know the specifics, I have a DB25 to DB9 converter, a homemade null printer cable - male to female DB9, and a USB to DB9.  The homemade cable simply has pins 2 and 3 crossed, with pin 5 connected - three wires in all.  Of course, this won't allow for any sort of flow control, but I'm cool with that.  You could also wire up every pin.  Check wikipedia for the pinout.  Wikipedia - Null Modem

 

TG

0 Kudos
Message 15 of 16
(3,453 Views)

Dennis,

 

Just one of the quirks of the Yokogawa WT200 series.  The LV libs from Yokogawa are written to comply with the 488.2 Standard.  A painful lesson I learned a long time ago and never forgot. 