Executing "TriggerList" using GPIB Interactive Control

Hi all,

 

 

For debugging purposes, I am trying to execute a multi-device "TriggerList" command using the "GPIB interactive control" utility.

 

I have a GPIB-USB-HS unit connected to 2 GPIB devices at addresses 11 and 12. Communication to these devices seperately works fine.

 

What I tried so far:

 

(1) Opened NI-MAX, browsed to "Devices and Interfaces" / "GPIB-USB-HS "GPIB0";

(2) Pressed the 'interactive control' button, which opens up the GPIB Interactive Control terminal window, with a "gpib0:" prompt;

(3) Enter "TriggerList 11,12" and press enter.

 

the result of this is:

 

[8000] (err)

error: EHDL

count: 0

 

Using the "NI I/O Trace" tool, I see that a TriggerList(31000, {0x000B, 0x000C}) call is executed, and returns with an EHDL error.

 

 

My question: what am I doing wrong, and what do I need to do to execute a successful TriggerList?

 

 

 

0 Kudos
Message 1 of 2
(4,280 Views)

Answering my own question:

 

The trick is to engage 488.2 mode. After that, triggering will work as intended:

 

    gpib0: set 488.2 0

    488.2 (0): SendIFC
    [0130]   ( cmpl cic atn )

    488.2 (0): TriggerList 11,12
    [0138]   ( cmpl cic atn tacs )
    count:  5

 

Monitoring the GPIB shows that this sends the 5-byte sequence  0x40, 0x3f, 0x2b, 0x2c, 0x08.

 