10-23-2015 05:38 AM
Hello.
I use Kiethley 2400 Sourcemeter which connected to PXIe-8135 by GPIB and PXIe-4139. I use real-time module on it.
I can see Kiethley from Visa Ineractive Control in localhost and in PXIe.
When I launch LabView program from the host computer I can see VISA resources, but can't see my Real-Time device PXIe-4139. When I launch it from Real-Time target, I can see PXIe-4139, but I can't see VISA resources.
I've tried to do so http://digital.ni.com/public.nsf/allkb/01F9536A7A1D12A486256FF000561DA6 , but it hadn't help.
How can I access both devices from LabView?
10-23-2015 06:40 AM
Did you install VISA on your Real Time target?
10-23-2015 06:56 AM
Yes
10-23-2015 08:19 AM
I solved my problem.
The problem was because I entered adress of Kiethley like this: visa://NI-PXIe8135.../GPIB0::24::INSTR but when launch LabView program from Real-Time system I should write adress like this: GPIB0::24::INSTR.
But this adresses in my program I writed directly, because in the drop down list were not anything except "Refresh" and it confused me.
The similar problem described here: http://forums.ni.com/t5/Instrument-Control-GPIB-Serial/VISA-Access-to-the-resource-or-remote-machine...