Hello,
I am trying to use Labwindows CVI 2013 with a GPIB card, Agilent 82351a and it can't talk with a GPIB instrument connected to the GPIB card.
NI-MAX can see the 82351a card but I can't use this card to talk with an instrument.
I installed Labwindows CVI with NI-VISA 5.4, AGILENT IO suite 16.3 with Agilent Visa a secondary VISA.
With NI-MAX, in the options VISA, I enable "Passport for Tulip" and with Agilent Connexion Expert, I tried to enable '"programs Agilent GPIB cards for 488" but it reboots the PC and the option remains not validated.
I wan't to know if this Agilent 82351a card can work with Labwindows CVI 2013.
thanks
04-22-2014 03:05 AM
Hello Benjamen,
According the readme file for NI VISA driver, you can use Agilent hardware with these steps :
NI-VISA now works with Agilent/HP GPIB and VXI controller cards. To use this feature, go into MAX, select Tools»NI-VISA»VISA Options, choose the Passports tab, and enable the NiVisaTulip.dll Passport. NI-VISA automatically finds and accesses any GPIB or VXI devices. However, NI-VISA does not configure the Agilent/HP hardware; you still must do that with the Agilent/HP I/O Config utility.
If you have both National Instruments and Agilent/HP controllers in the same system, you must configure the cards at different system addresses, such as GPIB0 and GPIB1. If both vendors try to use the same interface address, NI-VISA reports an error from viFindRsrc or VISA Find Resources."
Aslo be sure to use the last VISA version which is the 5.4.1
Thanks
Paolo_P
Certified TestStand Architect
Certified LabVIEW Architect
National Instruments France