420 Query-unterminated

when I try to communicate to Keithley Model 2000 by GPIB-RS232,I send "*IDN?\n" it is OK. When I send another command like ":CONFigure:FRESistance\n" ,the device show " ERROR -420 Query unterminated". I don't know why.Tell me!Thanks!

0 Kudos
Message 1 of 2
(2,877 Views)

@HelloWorld2 wrote:

When I send another command like ":CONFigure:FRESistance\n" ,the device show " ERROR -420 Query unterminated".


If you are trying to ask the instrument for what it is set to, you need a "?".  So your command really needs to be ":CONFigure:FRESistance?\n"


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5