myRio analog signal in LabVIEW

Hi all,

 

I am currently working on acquiring a 40kHz frequency through A/AI0 channel of the myRio board.  With LabVIEW, I am able to acquire half of the waveform.  Only the positive side of the waveform is shown for some reason.  I was wondering if there is a setting in LabVIEW to change this problem or in myRio.  I should be getting a -1V to +1V peak-to-peak waveform.  Instead, I am getting only the waveform from 0 to +1V and it is clipped at 0.  

 

Thank you for your help! 

0 Kudos
Message 1 of 4
(4,950 Views)

Hi pier2593,

 

That is expected behavior, because the AI inputs on the MXP A Connector is uni-polar.  This is documented in the specifications for your device (Table 1 on page 5 of the link below)

 

http://www.ni.com/pdf/manuals/376047a.pdf

 

Here is a screenshot for convenience:

myRIO_AI_problem.JPG

 

In summary: The channel you selected can only measure 0-5V.

 

If you need to measure a -1V to 1V signal, you will need to use the AI0/AI1 on the MSP (Connector C) on the myRIO.  These two AI inputs support +-10V.

 

myrio_ai2.JPG

 

I hope this helps and good luck!

 

Regards,

Nathan P.

Systems Engineer
SISU
0 Kudos
Message 2 of 4
(4,949 Views)

How would you go about using the +/- differential analog input channels?  For the moment, I am simply sending a 40kHz signal to the analog input C/AI0+ and C/AI0-.  Then I am trying to get both side of the analog signal without any good results.