6.5 MHz Resonant signal in 5752

The terminal interface is just a card with screw terminals to interface with the 68 pin connector.  It is true that we are using the 5752 in a single-ended mode by connecting the - to the ground. As for the impedance mismatch I can get a signal directly with the array that has as an impedance of 12 Ohm in the resistance and -130.5 Ohm in the reactance. I would think that this would be more of a mismatch than the 50 Ohm to 100 Ohm. We have also tried matching by using an inductor resistor circuit in addition to a 100 Ohm resistor between the + and ground (as the draw down resistor). Of which each was tested separately, but with no improvement to the signal. As for the harmonics in the signal I have attached an FFT of the signal. Looking at the spectrum it shows the 6.5 MHZ as the dominate frequency, with the 2.25 MHz more of a sub harmonic. Is it not true if the 6.5 MHz was a harmonic then the 2.25 MHz would be the dominate frequency.

 

 

I apperciate your input and help.

avantibob
0 Kudos
Message 11 of 18
(4,032 Views)

Are you sure that plot is correct, because if it is, then you don't have a 6.5MHz resonant signal, its a ~650kHz resonant signal...

Systems Engineer
SISU
0 Kudos
Message 12 of 18
(3,999 Views)

I am sure it is MHz there is an error in my Matlab code on the scaling of the x-axis, I had just missed it. 

avantibob
0 Kudos
Message 13 of 18
(3,995 Views)

Just to clarify, the image you attached indicates:

  • There is a fundamental ~335kHz tone (not MHz)
  • There is a 2nd order harmonic (not 3rd) of that fundamental at about 671kHz (not MHz)
  • 3rd order harmonics, and all other harmonics (not even, not odd, but all) are also present.

The amplitude flatness of the 5752 below about 1MHz degrades, and this changes somewhat with the configuration of the ADC.  If you have a fundamental resonance at ~335kHz (e.g. due to cable lengths from your source, through switches, to the FlexRIO Adapter Module), then it's feasible that measured amplitude of harmonics vary from (are greater or less than) the ~335kHz fundamental.

 

General questions:

From your signal source to your 5752, can you estimate for the total cable length (cables/wires/switches/etc)?

 

Do you have access to a time domain reflectometer?  I believe that would be the best instrument to use in order to determine if there are impedance mismatches along any of the path length of the signal to your 5752 that are causing a ~335kHz resonance.  A TDR will provide you a plot of characteristic impedance (in ohms) vs. distance, so you can characterize what your signal sees as it goes through your system's cables, switches, etc.).  This will help you determine at which point in your signal path you may have a temporary shift in impedance (an impedance mismatch), which would cause reflections at a particular frequency.

 

-Andrew

National Instruments
0 Kudos
Message 14 of 18
(3,994 Views)

Sorry for any confusion but there is an error in the scaling of the x-axis

the fundamental would then be 3.35 Mhz

the 2nd 6.71 MHz. I know because I counted 6 cycles in 1 msec of the signal in the time domain.

I can estimate that there approximately 1.5 meters of distance from the source.

Unfortunately I don’t have a domain reflectometer. Is there another way to test this?

 

Thanks for your help

avantibob
0 Kudos
Message 15 of 18
(3,986 Views)

6 cycles in 1ms... 

 

6 cycles/1ms * 1000ms/1s = 6000 cycles/sec = 6kHz.

 

Are you sure your scales and math are correct?

Systems Engineer
SISU
0 Kudos
Message 16 of 18
(3,949 Views)

It looks like I can't type as fast has I has thinking and got a little frustrated because the pin drive with the data took a dump so I could not correct the scaling of the FFT properly. I counted 6 cycles in 1 micro second

 

6 cycles/1 microsecond * 1000000 microseconds/1sec= 6MHz

 

Again I apologize for the confusion. I am trying to get a hold of the TDR to check the cables.

avantibob
0 Kudos
Message 17 of 18
(3,945 Views)

It is possible that your preamp (30bd gain!)  is oscillating. Scope with input 50Ohm impedance is fine but 100Ohm from the 5752 can cause reflections leading to oscillation.

I would test with differential input shunted with an additional 100Ohm SMD resistor (rigth at the input)  and a 1Meg (to 10k)  from AI- to AIGND to provide a bias path.

Some amplifiers a sensible to capacitive loads , thats why impedance matching and short cables are important.

What will happen on the scope if you add another 50Ohm (maybe build from two 100Ohm SMD or a 49Ohm) in series at the scope input leading to a 100Ohm missmatch input impedance? 

 

And use SMD resistors

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy