02-03-2016 10:18 AM
The FP-TB-3 has a CJ imbeded into the TB. This is not calibrated separatly but is used in conjuction with FP-120. The FP-120 is calibrated for CJC but is never linked to the TB. There should be a system calibration between the TB and FP-120 and be used togheter?
Do you know the thermistor type used in the FP-TB-3?
There is a posibility to measre the thermistor propertise from the terminals?
02-04-2016 02:13 PM - last edited on 03-05-2024 10:06 AM by migration-bot
Hi tfulop,
Just in case, have you taken a look at the FP-TC-120 and cFP-TC-120 Operating Instructions manual?
The document itself can be found here:
https://www.ni.com/docs/en-US/bundle/373344a/resource/373344a.pdf
02-05-2016 03:17 AM - last edited on 03-05-2024 10:07 AM by migration-bot
Does this help?
There are three different thermistors that are currently being used in our FieldPoint terminal bases and are recommended for CJC.
The thermistor needs to be connected between pins 36 and 37 in order to function properly.
Model Part Number
U.S. Sensor PT502J2
BetaTHERM 5K3A1
Sensor Scientific WM502C
Each of these thermistors are interchangeable, negative temperature coefficient (NTC) thermistors with 5kohm @ 25 C, +/- .2 C. Specifications for the thermistors can be found at the manufacturers website.
The important characteristic of these thermistors is that the resistance vs. temperature ( R vs. T) curve is similar between them. Also, each thermistor contains a beta value for the R vs. T curve, which produces similar results across the specified temperature range. Although different thermistors may contain closely matching beta values, the R vs. T curve may be different and may not perform as expected.
Cheers
Stephen
02-05-2016 08:23 AM
Thanks, right what I was looking for.