Hello Harley-Alex,
Alan From LeCroy Tech Support here. The various Read Waveform VIs execute the "Arm" command, so it isn't for use in Auto or Normal trigger mode. We recommend using this command since it utilizes Single trigger mode, which is the mode to use when you want your program to control when the scope triggers, and thus lets you be sure that you are reading back data that corresponds to a specific trigger event.
If you prefer to run in Auto or Normal mode, you should use the "Wait For Acquisition Complete" VI that is in
the Data/Low Level function palette. The looks for the "INR" register to show evidence of a new acquisition.
Best Regards,
Alan