10-29-2013 10:55 AM
I am concerned about the temp. of the 4461 card in my PC, the specification of the card says that the normal operating temp. is 0 to 50C, when i checked the temp. on MAX under the current temp. It's 49C! my room temp. is around 22C, i also assume that when the card starts working and generating signals the card temp. will go even higher! any thoughts on this please.
Omran
10-30-2013 10:26 AM
Hello techtak
That information MAX displays is not the current temperature, it is the last self calibration temperature.
Self Calibration >> Last Self Calibration Temperature
Your card might be working under 50C, you could perform another self calibration at a lower temp if you want.
Regards
Frank R.
10-30-2013 10:30 AM
I would also suggest cleaning your fan filters if you have them.
10-31-2013 09:27 AM
Hello Frank,
Thanks for taking the time on this, in fact there is a current temp. monitoring in MAX as you see the attached pic. and the temp. does change depending on the environment around, but I have to restart MAX every time to see the change in temp.! i opened the PC cover and directed a small cooling fan on the card and the temp. dropped down to 40C in few minutes, with the cover open and without a fan the temp only went down to 47 from 49!
10-31-2013 09:39 AM
Hello techtak
You are right, that one is the current temp. You should assure its functionality under 50C.
Getting the Current Temperature in Measurement & Automation Explorer (MAX) for Your Device.
Regards
Frank R.
11-01-2013 10:15 AM
In MAX, the only thing you should need to do to update the current temp is hit the Refresh button in the toolbar. I test this using MAX 5.5 and DAQmx 9.8 as a reference.
11-01-2013 11:09 AM
Thanks Joshua, I also realized that I can refresh by hitting F5,