Digital Multimeters (DMMs) and Precision DC Sources
Trigger timeout PXI DMM 4070
Hello
I'm using a DMM PXI-4070 and the DotNet example "SoftwareTriggeredMultipointAcquisition".
I want to trigger the DMM4070 with the AuxTrig input on the front of the instrument.
This works fine but how can i change the time out the of triigger ?
This is now set to 2000msec. i want to set this to at 10000msec.
When i us the example, and do not trigger within 2000msec i get an error message saying :
ModularInstruments.NIDmm: The operation did not complete within the maximum time allowed. Timeout: 2000mS.
Error Code : -1074126845
Thanks in advance
Tonnie
(5,051 Views)
The timeout is connected to the DMMMeasurement.FetchMultiPoint method.  The version of the method in the example just uses the pointsToFetch input parameter.  You're currently seeing the default timeout value. You can add a timeout input parameter to this function in the code.
 The overload list for the method is below:
The syntax for the version of the method you'll need to use is below:
Let me know if you run into any problems getting this done.
Jared R.
Precision DC Product Support Engineer
(5,040 Views)
Accepted by topic author Tonnie
Hello JaredRo
I can set the timeout now.
Problem solved. Manny thank
With kind regards
Tonnie
(5,032 Views)
