DMM Multipoint Software trigger works as immediate

HI,

 

We are using NI PXI 4071 DMM in .NET (C#) and see a strange behavior of the instrument set for multipoint acquisition with software trigger.

We are trying to acquire multiple series (let's say 5), where each set having 4 samples,

Our settings are

 TriggerSource = nidmmConstants.SoftwareTrig;

 SampleTrigger= nidmmConstants.SoftwareTrig;

 

and we sending calls SendSoftwareTrigger() for each sample;

 

1. We see that acquisition of the first sample in each series is not waiting for the SendSoftwareTrigger() call but done immediately!

Total number of the SendSoftwareTrigger() calls required for the data acquisition is not 20 but 15!

 

2. How we can setup start of acquisition of the each set by software trigger, but acquisition of the each sample in the series immediately with time interval?

So, for 5 series we will send only 5 SW triggers, where each one should start acquisition of 4 samples one after another till next set?

 

Expected combination of the

   TriggerSource = nidmmConstants.SoftwareTrig;

   SampleTrigger= nidmmConstants.Immediate;

does not work. We have seen immediate acquisition of the all 20 samples!

0 Kudos
Message 1 of 20
(8,032 Views)

Hi gremlm,

 

Can you share a bit more of your code to help us understand how you configure and start your acquisition tasks?  Am I correct in understanding that you'd like to acquire 4 samples from the same line immediately after a software trigger, wait some time, then acquire another 4 samples after another trigger, and so on?

 

Have you taken a look at any of the text-based examples provided with the DAQmx drivers?  I realize you are using NI-DMM, but the triggering should be similar.

 

0 Kudos
Message 2 of 20
(8,011 Views)

Our problem is that even when we set up TriggerSource to software trigger, data aquisition happened immeditedly

By the way we are familiar with DAQmx very well, there are no such problems there.

 

I think that there is an error in NI 4071 DMM driver.

 

Because if we set the TriggerSource first to External trigger and than to the Software, everything works exactly as we expecting!

Transition of the TriggerSource from Immedite to Software trigger directly does not work!

Transition of the TriggerSource from Immedite External than to Software trigger works!

 

We have tried to use a digital line for triggereing (we have a PXI  system and used PXI trigger line #5).

Exactly the same problem was detected:

Transition of the TriggerSource from Immedite to TTL5 trigger directly does not work!

Transition of the TriggerSource from Immedite External than to TTL5 trigger works!

 

Another confusing issue:

We TriggerSource  to Software trigger and reset device through the Reset command

Than we check the the value of the TriggerSource.

It is not restored back to Immedite, but is equal to the Software trigger!

 

There is a definetly an error in NI 4071 DMM driver!

 

0 Kudos
Message 3 of 20
(8,003 Views)

Hey Gremlm,

 

What happens if you change the trigger from Immediate to TTL5, then to TTL5 again?

Can you post the relevant section of your code, or let us know which example you're using?

 

...

 

Which reset command are you calling? (niDMM Reset or niDMM Reset with Defaults)

 

-John Sullivan
Problem Solver
0 Kudos
Message 4 of 20
(7,983 Views)

>>What happens if you change the trigger from Immediate to TTL5, then to TTL5 again...

 

What is a point to set up trigger to TTL5 multiple times? It doesn't work from the first one?

If we sets TriggerSource to External than to any other - it always works!

Question is why it doesn't work directly!

 

For reset  we has 

      niDriver.reset();

 

Should  we execute 

     niDriver.ResetWithDefaults() instead?

 

 

 

0 Kudos
Message 5 of 20
(7,978 Views)

Hi gremlm,

 

Is it possible for you to post some code that shows the issues you are having?  So far I have tried replicating the incorrect triggering behavior that you're describing without success.  Being able to work with your code would be helpful.  Are you using niDMM Read or niDMM Fetch to return your samples?

 

What version of NI-DMM are you using?  What values are you using for Trigger Count and Sample Count?  If you are taking 5 sets of 4 measurements, your trigger count should be 5 and your sample count should be 4.  These are the values that I used.

 

When I set TriggerSource to SoftwareTrig and SampleTrigger to SoftwareTrig, no samples are returned until I send a software trigger.  Using this configuration, one sample is returned for every software trigger I assert.  

 

When I set TriggerSource to SoftwareTrig and SampleTrigger to Immediate, no samples are returned until I send a software trigger.  Using this configuration, 4 samples are returned for every software trigger I assert.  

 

 

I was able to reproduce the behavior you described regarding niDMM Reset.  TriggerSource is not reset to its default value when Reset is called.  You should explicitly set the TriggerSource at the begininng of your code to get around this issue.

 

Let me know the answers to these questions and we can continue troubleshooting.

 

Regards,

Jared R.
Precision DC Product Support Engineer
National Instruments
Message 6 of 20
(7,960 Views)

Hey GremIm,

 

>>What is a point to set up trigger to TTL5 multiple times?

My recommendation was to determine if there was a user-related programming issue.  We still don't know if you're using NI's example code, or your own custom configuration.  If you're able to reproduce the issue with NI-published examples, that will help us hone in on the answer.  If you're using your own code, post the relevant section here and we can see if there's incorrect API usage.  I agree that the code should be capable of working directly, without the intermediate call to external trigger, but we first need to reproduce the behavior you're seeing here, so we can see where the issue resides.

 

Have a great day!

-John Sullivan
Problem Solver
0 Kudos
Message 7 of 20
(7,954 Views)

Do you have an example where Software trigger is used for 4071 DMM multipoint data acquisition under .NET so we can try it?

 

About our code - it is a rather a big application, but here is simplified version what we are doing (C# code:)

 

 

EXAMPLE1

// Setting measurement function DCVoltage, range,  presicion,  aperture, etc.

 

 

niDriver.SetInt32(nidmmProperties.TriggerCount, 1);

niDriver.SetInt32(nidmmProperties.SampleCount, 10);

 

nidmmConstants.External

niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.SoftwareTrig);

niDriver.SetInt32(nidmmProperties.SampleTrigger, nidmmConstants.SoftwareTrig);

 

niDriver.Initiate();

 

//Commented:

// for(int i=0; i< 10; i++)

//{

//    niDriver.SendSoftwareTrigger();

//    Thread.Sleep(200);

//}

 

Double[] result = new Double[10];
int numberOfPoints = 0;

niDriver.FetchMultiPoint( 1000, 10, result,  out numberOfPoints);

 

and data is aquired without sending software triggers!

but

EXAMPLE1.1

 

// Setting measurement function DCVoltage, range,  presicion,  aperture, etc.

 

niDriver.SetInt32(nidmmProperties.TriggerCount, 1);

niDriver.SetInt32(nidmmProperties.SampleCount, 10);

 

niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.External);

niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.SoftwareTrig);

 

niDriver.SetInt32(nidmmProperties.SampleTrigger, nidmmConstants.SoftwareTrig);

 

niDriver.Initiate();

 

//Commented loop:

// for(int i=0; i< 10; i++)

//{

//    niDriver.SendSoftwareTrigger();

//    Thread.Sleep(200);

//}

 

Double[] result = new Double[10];
int numberOfPoints = 0;

niDriver.FetchMultiPoint( 10000, 10, result,  out numberOfPoints);

 

data not aquired without sending software triggers!

It only works if "Commented loop:...." code is un commented!

 

The same is true for the following code:

EXAMPLE2

 

// Setting measurement function DCVoltage, range,  presicion,  aperture, etc.

 

niDriver.SetInt32(nidmmProperties.TriggerCount, 1);

niDriver.SetInt32(nidmmProperties.SampleCount, 10);

 

niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.SoftwareTrig);

niDriver.SetInt32(nidmmProperties.SampleTrigger, nidmmConstants.Immediate);

 

niDriver.Initiate();

 

// niDriver.SendSoftwareTrigger();

 

Double[] result = new Double[10];
int numberOfPoints = 0;

niDriver.FetchMultiPoint( 10000, 10, result,  out numberOfPoints);

 

and data is aquired without sending software trigger for entire set!

but

EXAMPLE2.1

 

 

// Setting measurement function DCVoltage, range,  presicion,  aperture, etc.

 

niDriver.SetInt32(nidmmProperties.TriggerCount, 1);

niDriver.SetInt32(nidmmProperties.SampleCount, 10);

 

niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.External);

niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.SoftwareTrig);

 

niDriver.SetInt32(nidmmProperties.SampleTrigger, nidmmConstants.Immediate);

 

niDriver.Initiate();

 

// niDriver.SendSoftwareTrigger();

 

Double[] result = new Double[10];
int numberOfPoints = 0;

niDriver.FetchMultiPoint( 10000, 10, result,  out numberOfPoints);

 

and data is not aquired without sending software trigger.

It only is acquired if

                  niDriver.SendSoftwareTrigger();

is uncommented!

 

 

 

0 Kudos
Message 8 of 20
(7,946 Views)

The NI DMM version is 3.0.4

0 Kudos
Message 9 of 20
(7,940 Views)

Hi gremlm,

 

Sorry for the delayed response.  I set up a test machine and ran the code you provided.  I wasn't able to replicate any of the problems you described.  

 

When I ran Example 1 and Example 1.1, I do not recieve data unless the for loop is uncommented.  If the for loop is commented out, I receive a timeout error, as expected.

 

When I ran Example 2 and 2.1, I get a timeout error unless I uncomment the software trigger you sent, in which case all 10 samples are returned. 

 

I am not sure why your system seems to be generating false software triggers.  You mentioned that is a big application.  Is is possible that software triggers are being sent from elsewhere in your code?

 

The program that I was running is below:

using System;
using System.Collections.Generic;
using System.Linq;
using System.Text;
using System.Threading.Tasks;
using NationalInstruments.ModularInstruments.Interop;

namespace DMM
{
     class Program
     {
          static void Main(string[] args)
          {
          nidmm niDriver;

 

          niDriver = new nidmm("PXI1Slot8", false, false);

 

          niDriver.SetInt32(nidmmProperties.TriggerCount, 1);
          niDriver.SetInt32(nidmmProperties.SampleCount, 10);

 

          niDriver.SetInt32(nidmmProperties.TriggerSource, nidmmConstants.SoftwareTrig);
          niDriver.SetInt32(nidmmProperties.SampleTrigger, nidmmConstants.SoftwareTrig);

 

          niDriver.Initiate();

 

          //for (int i = 0; i < 10; i++)
          //{
          //      niDriver.SendSoftwareTrigger();
          //      System.Threading.Thread.Sleep(200);
          //}

          Double[] result = new Double[10];
          int numberOfPoints = 0;
          niDriver.FetchMultiPoint(1000, 10, result, out numberOfPoints);

          var builder = new StringBuilder();
          Array.ForEach(result, x => builder.Append(x + "\n"));
          var res = builder.ToString(); //builds string from DMM results

 

          Console.WriteLine(res + "\n" + "Number of Points: " + numberOfPoints); //Prints data to command line window
          System.Console.Read(); //Waits for user input

 

          niDriver.Abort();

          

          }
     }
}

Regards,

Jared R.
Precision DC Product Support Engineer
National Instruments