Calculating PXIe-4145 SMU Resolution

Solved!
Go to solution

I am trying to calculate the resolution of my PXIe-4145 4CH Precision SMU, but Figure 2 in the PXIe-4145 Spec Sheet appears to have inconsistent labels on the x-axis. If all labels were in msec (not mixed usec/msec), it would map to a consistent logarithmic scale as the graph implies.

 

PXIe-4145 SMU Resolution graph.jpg

 

aperture (sec) resolution (ppm) notes
0.010000000 1 if graph didn't curve
0.001000000 10  
0.000100000 100 incorrectly labeled as 0.1 usec
0.000010000 1000 incorrectly labeled as 0.01 usec
0.000001667 6000 I think this is the max sample rate of 600kS/sec
0.000001000 10000 if graph continued

 

link to spec sheet:

http://www.ni.com/documentation/en/pxi-source-measure-unit/latest/specs-pxie-4145/specs/#GUID-D04476...

 

Message 1 of 3
(4,338 Views)
Solution
Accepted by topic author sam.broyles

Yes, you're correct; the X-axis labels should be 1µ, 10µ, 100µ, 1m, 10m, and 100m.  Thanks for pointing this out.

Message 2 of 3
(4,277 Views)

Hey Sam, 

 

Thank you for bringing this to our attention. 

I've filed a Corrective Action Request (CAR) to fix this in the 4145 specs (and also the 4144) specs.

 

Cheers, 

Chris J.

Product Support Engineer

NI

Message 3 of 3
(4,270 Views)