Hi Else,
there is no correct setting for FFT calculations. You can get a lot of
different results and most of them are making sense. On the first view, this
might look very complicated, but most of the time, the settings are only
important if you try to compare your results with the results of other FFT
Analysers or software. If you just need to find out more about your data, you
can try out different settings and look for the results you need.
Here are some answers to your questions.
Your signal is cut of at the begin and the end of the measurement which results
in an error. Most of the window functions are reducing this error. Hanning is
the most common window. You should compare Hanning and Rectangle (no window) to
see which one is giving you the better result. Forget all the other windows if
you don't know them.
Hanning is taking a part of your signal away. In the first place, all
FFT-values are smaller then the results without a window. It is possible to
multiply the results with a correction. If you are looking for a sinus wave
within your signal and you are interested in the amplitude of this sinus, you
should use the periodic correction. In all other cases you should use the
random correction. But the correction is just a factor. If you are not interested
in the absolute value of the amplitudes at all, it doesn't matter.
FFT functions:
It is always the same result, but sometimes squared or multiplied with a
factor.
Amplitude is OK, unless you compare your values with other results.
RMS = Amplitude/square root(2)
Autospectrum = Amplitude^2
Powerspectrum = RMS^2
Averaging is only important if you do more than one FFT calculation. In the
first place, I would do one FFT as large as possible. If your signal is a continuous
signal, you get better results if you take as much values as possible (4096,
8192, ...). Larger FFTs are giving you a much better frequency resolution. Most
of the averaging and overlapping is not useful anymore. We have so much memory
and speed in our computers, that we can do calculations with millions of
values. Today averaging is still very important for 2-channel calculations like
transfer functions on numbers of short measurements.
DIAdem can handle datasets with gaps where the equipment was not able to
deliver any value. "Reject no values" is not important for you,
because you don’t have "no values" in your measurement.
If 0 Hz is the dominant frequency, using a high pass filter is a good way to
get better results.
This are just some short answers to your questions. Signal processing and FFT is a very wide area with lots of things you might need to know. If you need more specific help on analysing your data you can post a set of these data or send it by mail. It also would help to know what you are trying to find out.
regards,