Hello all
I am trying to write a pulsetrain to the counter output of a PXI-6624, however only the first sample is ever written.
My pulsetrain has 50 samples, and with simulated hardware the output "number of samples written per channel" matches that. If I build the application and run the program on the target system it only outputs one sample indefinitely and also "number of samples written per channel" is always 1.
I tried adjusting the value for "numSamplesWriten" in "DAQmx Write (Counter 1D Time 1Chan NSamp).vi" to a non-zero value without any influence. The timing is set to 50 samples in continuous sample mode with implicit timing.
What am I missing, how do I get the counter to go through the array of samples one by one to output the desired timing series?
Thanks in advance,
Raphael
Solved! Go to Solution.
04-04-2017 11:03 AM
I think the 6624 is an older model with an NI-TIO timing chip that doesn't support buffered pulse train output. The best you can do is to change on-the-fly with software timing.
Note also that finite pulse trains will occupy a 2nd counter known as the "paired counter". The pairing is set internally, cannot be changed, and goes like:
0<-->1
2<-->3
4<-->5
6<-->7
A newer card like the 6612 can do buffered pulse train output with hardware-level timing, *and* finite pulse trains no longer occupy a 2nd paired counter.
-Kevin P
04-12-2017 02:30 AM
Hello Kevin
thank you for your reply. It really seems it is not possible with this hardware.
Since the hardware is already fixed I will have to think of something else.
Thanks, Raphael