Are you able to communicate with the device using Measurement and Automation Explorer (MAX)? Visit the
GPIB Installation/Configuration Troubleshooter for instructions on how to interactively control your device.
You can use the GPIB Analyzer to capture/monitor bus activity. In MAX, select Tools » NI-488.2 » GPIB Analyzer. The analyzer main window appears, containing the Bus Monitor window, the Action window, and a new Capture Display window.
Other helpful GPIB tips:
When using GPIB, it is very important that the correct driver version is installed for the operating system you are running. Click
here to view a table of t
he current driver versions by board and operating system.
The maximum separation between any two devices is 4 m and the average separation cannot be greater than 2 m over the entire bus.
A maximum of 15 loads/devices per bus (e.g., one board and up to 14 instruments) is allowed and at least two-thirds of the devices must be powered on.
If these suggestions do not solve your problem, please be more specific about the error message you are receiving. A screenshot would help clarify the situation. Have a great day!